Quick links
- "No Radioactive Material Use Report" form
- Radioactive Material Order Request" form
- Radiation Dosimeters and Radioactive Waste Disposal Prices for FY10 (pdf)
- Environmental Safety Facility - for chemical concerns at UVM
- Understanding your Annual Radiation Exposure Report
- Policy for minors working in a radiation lab (pdf)
Related links
- Perkin Elmer -radiochemicals
- MP (formerly ICN) - radiochemcials
- Ludlum Measurements, Inc. - portable radiation detectors
- NRC -Nuclear Regulatory Commission
- EPA - Radiation Protection
- FDA - Center for X-ray Device and Radiological Health
- Health Physics Society - "Ask The Experts"
- Radiation Answers - Learn more about radiation
STANDARD OPERATING PROCEDURES
Phosphorus-33
version: December 10, 1996
INTRODUCTION
Phosphorus-33 is becoming more commonly used radionuclide with a half-life of 25.4 days, emitting only beta particles with a maximum energy of 0.249 MeV (Million Electron Volts) and an average energy of 0.083 MeV. The beta particles from P-33 travel a maximum of 46 cm in air.CONCERNS
The major concern with using P-33 is that it cannot be easily monitored during it's use, therefore, special precautions are needed to keep the work environment clean. The regular use of wipe testing is the only way to insure that your work space is not contaminated. Contamination on the skin will not likely cause a significant dose to the dead layer of skin, however, it could lead to the internal absorption of P-33. The maximum permissible body burden to the whole body is 3 millicurie.SHIELDING
Glass and plastic are the best shields for beta particles from P-33.DETECTION
A tiny drop of contamination from P-33 can be easily detected with a wipe test from a Liquid Scintillation Counter. Most Geiger Counters will not efficiently detect the presence of P-33.EQUIPMENT / SUPPLIES
The following equipment and supplies must be available:- a Liquid Scintillation Detector.
- disposable latex or plastic gloves.
- a full-length lab coat.
- radioactive waste receptacle
- pipettes dedicated to the use of P-33.
- commercial decontaminate, i.e. DuPont's "Count Off".
SAFETY RULES
If the following safety precautions are used, personnel radiation exposure will be as low as reasonably achievable.- Designate a specific area of the lab for P-33 handling.
- Full-length lab coats must be worn by all persons who handle P-33.
- Protect your hands from becoming contaminated from spills by wearing two pairs of disposable gloves.
- Never pipette P-33 by mouth.
- Only use pipettes which have been dedicated to your specific use of P-33.
- Pipettes will easily become contaminated and therefore, should not be shared with others.
- If you have reason to believe that your gloves are contaminated, immediately dispose of them in the radioactive waste container
POST-USE PROCEDURES
After handling P-33,- Conduct a wipe test and count the wipes in a Liquid Scintillation counter:
- Check all equipment, centrifuges, water baths for
contamination.
- If any contamination is found, use a commercial radiation contamination remover (i.e. Count Off) with paper towels to clean up the equipment.
- Place the towels in the radioactive waste receptacle.
- If contamination cannot be removed, place a "radiation" label on the equipment indicating that it is P-33, maximum cpm found, and the date you measured the level.
- Check the work bench and floor.
- If contamination is found, it can usually be removed easily with "Count Off". If it cannot be removed, contact the RSO to obtain shielding materials.
- Inform your fellow lab workers if any unremovable contamination is found.
- Check the normal trash container to make sure no radioactive waste has been accidentally placed there.
- Store the waste temporarily in containers marked with labels "Radioactive Waste-Do Not Empty". These labels are available in the RSO.
- Send a Radiation Contamination Survey
Report to the RSO.
- Call the RSO if you have any questions about where to survey, or how to fill out the form.
- Wash your hands thoroughly.
- Bring the waste frequently to the RSO. We accept waste
every Friday. Call 656-2570 to make an appointment.
º Any questions about these procedures ? Call the Radiation Safety Office, 656-2570 ºLast modified January 21 2003 09:34 AM
